metcal.ini port statement

Hello,

I'm using both METCAL 7.3 and 8.1. What should I edit in the metcal.ini "port = ...?" statement to make the computer accept any calibrator, 5500, 5520, 5800 and/or 9500? I want interchangeability, not getting stuck with only one type of standard? My computer seems to be married to the 9500 calibrator when set at port = COM3 and will not accept a 5520! On another computer, it's gladly accept a 5522A if I set port = M5500 and refused to work at all on COM2, or COM3! Any answer to this problem?

I thought this is a port statement in the .ini file, but what else could there be? Please help.

Thanks.

3 comments

Date Votes
0
Avatar
Dale Chaudiere

1. The MET/CAL Initialization file parameter "port" set the Default Serial Port for RS-232 controlled DUTs.

2. The 9500 is IEEE-488 only.  It has no RS-232 port.  Check the rear panel.

3. Only the 5500A, 5502A, 5502E, 5520A, 5522A, 5800A, and 5820A have a "Serial 2 to UUT" DB9 RS-232 port on their rear panel.

4. The "Default Serial Port" CAN be set to this "Pass-through" RS-232 port on the rear panel of the Calibrator, when the Calibrator is configured.

5500A => COM5500

5502A => COM5500A

5502E => COM5502E

5520A => COM5520

5522A => COM5522

5800A => COM5800

5820A => COM5820

Whatever problem you are having setting the "Default Serial Port" has nothing to do with configuring a 9500.

0
Avatar
Tam Tu

Dale,

Where should I look in the .ini file to edit for allowing standards change as needed? Right now, this particular computer only let me use the 9500 but wont take 5520A? Thanks.

0
Avatar
Dale Chaudiere

References (System Instruments, User-configured Instruments, and Manual Standards) are not specified in the MET/CAL initialization file.  Configure references using the menu item in the Editor or Run Time application.  This depends upon what version of MET/CAL you are using.  However, F12 should open in the "Instrument Configuration" in ALL versions of MET/CAL.