0

Fluke 45

I make a procedure with a fluke 45.

But when i want to run the system says he can't find the accuracy file for Fluke 45.

Can someone help me what i do wrong.

this the procedure

7.005   45                 100.00Z                                          S 2W

7.006   MEMC           100.00Z           10U 

8 comments

Date Votes
0
Avatar
Héctor Alejandro Raimundo Cabrejo
Begin Header
instrument = Fluke 45
interval   = 6 months
confidence = 2.58 sigma
End Header
 
# $Revision: 3789 $
 
Mode: Volts
 
# DC
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
-99.999e-3  99.999e-3   NA    NA     0.02  6e-6    1e-6
100e-3      999.99e-3   NA    NA     0.02  60e-6   10e-6 
-999.99e-3  -100e-3     NA    NA     0.02  60e-6   10e-6 
1           9.9999      NA    NA     0.02  600e-6  100e-6
-9.9999     -1          NA    NA     0.02  600e-6  100e-6
10          99.999      NA    NA     0.02  6e-3    1e-3
-99.999     -10         NA    NA     0.02  6e-3    1e-3
100         999.99      NA    NA     0.02  60e-3   10e-3
-999.99     -100        NA    NA     0.02  60e-3   10e-3
 
# AC (15,000 counts to 99,999 counts, full range)
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
0          14.999e-3    20    100e3  NA    NA      1e-6
 
  15e-3      99.999e-3    10    19.99  NA    NA      1e-6
  15e-3      99.999e-3    20    50     1     100e-6  1e-6
  15e-3      99.999e-3    50    10e3   0.2   100e-6  1e-6
15e-3      99.999e-3    10e3  20e3   0.5   100e-6  1e-6
15e-3      99.999e-3    20e3  50e3   2     200e-6  1e-6
15e-3      99.999e-3    50e3  100e3  5     500e-6  1e-6
 
100e-3     999.99e-3    10    19.99  NA    NA      10e-6
100e-3     999.99e-3    20    50     1     1e-3    10e-6
100e-3     999.99e-3    50    10e3   0.2   1e-3    10e-6
100e-3     999.99e-3    10e3  20e3   0.5   1e-3    10e-6
100e-3     999.99e-3    20e3  50e3   2     2e-3    10e-6
100e-3     999.99e-3    50e3  100e3  5     5e-3    10e-6
 
1          9.9999       10    19.99  NA    NA      100e-6
1          9.9999       20    50     1     10e-3   100e-6
1          9.9999       50    10e3   0.2   10e-3   100e-6
1          9.9999       10e3  20e3   0.5   10e-3   100e-6
1          9.9999       20e3  50e3   2     20e-3   100e-6
1          9.9999       50e3  100e3  5     50e-3   100e-6
 
10         99.999       10    19.99  NA    NA      1e-3
10         99.999       20    50     1     100e-3  1e-3
10         99.999       50    10e3   0.2   100e-3  1e-3
10         99.999       10e3  20e3   0.5   100e-3  1e-3
10         99.999       20e3  50e3   2     200e-3  1e-3
10         99.999       50e3  100e3  5     500e-3  1e-3
 
100        750          10    19.99  NA    NA      10e-3
100        750          20    50     1     1       10e-3
100        750          50    10e3   0.2   1       10e-3
100        750          10e3  20e3   0.5   1       10e-3
100        750          20e3  50e3   2     2       10e-3
100        750          50e3  100e3  5     5       10e-3
 
Mode: Volts F
 
# 6kV probe
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
-6000      6000         NA    NA     1     200e-3  100e-3
 
Mode: Volts G
 
# 40 kV probe
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
-40e3      40e3         NA    NA     4     200e-3  100e-3
 
Mode: Amps
 
# 100mA Range
 
# DC
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
-9.9999e-3  9.9999e-3   NA    NA     0.05  1.5e-6  100e-9
10e-3       99.999e-3   NA    NA     0.05  5e-6    1e-6
-99.999e-3  -10e-3      NA    NA     0.05  5e-6    1e-6
 
# AC (15,000 to 99,999 counts, full range)
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
0           1.4999e-3   20    20e3   NA    NA      100e-9
 
1.5e-3      9.9999e-3   20    50     2     10e-6   100e-9
1.5e-3      9.9999e-3   50    10e3   0.5   10e-6   100e-9
1.5e-3      9.9999e-3   10e3  20e3   2     20e-6   100e-9
 
10e-3       99.999e-3   20    50     2     100e-6  1e-6
10e-3       99.999e-3   50    10e3   0.5   100e-6  1e-6
10e-3       99.999e-3   10e3  20e3   2     200e-6  1e-6
 
Mode: Amps E
 
# 10A Range
 
# DC
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
-10         10          NA    NA     0.2   700e-6  100e-6
 
# AC
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
0           0.5         20    2e3    NA    NA      100e-6
 
0.5         1           20    50     2     30e-3   100e-6
0.5         1           50    2e3    1     30e-3   100e-6
 
1           10          20    50     2     10e-3   100e-6
1           10          50    2e3    1     10e-3   100e-6
 
Mode: Ohms 2W
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
0           98          NA   NA      0.05  28e-3   1e-3
98          980         NA   NA      0.05  100e-3  10e-3
.98e3       9.8e3       NA   NA      0.05  800e-3  100e-3
9.8e3       98e3        NA   NA      0.05  8       1
98e3        980e3       NA   NA      0.06  80      10
.98e6       9.8e6       NA   NA      0.25  600     100
9.8e6       98e6        NA   NA      2     2e3     1e3
 
Mode: Hertz
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  ------
0           999.99      30e-3  750   0.05  20e-3   10e-3
1e3         9.9999e3    30e-3  750   0.05  100e-3  100e-3
10e3        99.999e3    30e-3  750   0.05  1       1
100e3       300.00e3    100e-3 750   0.05  10      10
300e3       999.99e3    1      750   0.05  10      10
 
Mode: Watts
 
# DC
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  -------
0.01       999.99       NA    NA     NA    NA      0.01
 
# AC
 
# NOMINAL               MOD1         Tol.  Floor   Res.
#---------------------  -----------  ----  ------  -------
0.01       999.99       20    50     2     0.1     0.01  
0.01       999.99       50    10e3   0.4   0.1     0.01  
0.01       999.99       10e3  20e3   1     0.1     0.01  
0.01       999.99       20e3  50e3   4     0.2     0.01  
0.01       999.99       50e3  100e3  10    0.5     0.01  
 
0
Avatar
Héctor Alejandro Raimundo Cabrejo

This is the Accuracy file that I have, you can compare with your acc file or you can copy paste in the directory

0
Avatar
Michael J.

Hi David,

It's always good to check the alt folder of your acc file location (C:\ProgramData\Fluke\METCAL\shared\acc\alt) first to make sure the missing accuracy file wasn't put there. When MET/CAL does its scan of the accuracy files, the alt folder which is contained inside the acc folder is not searched, so people often move accuracy files they don't need at the time into that folder to reduce the scan time at procedure launch.

If the Fluke 45 acc file isn't there, you can send an email to softwaresupport@flukecal.com and we can send you a current copy.

0
Avatar
David Van der Meersch

Hector,

Allready thanks for acc file.

I have checked the acc file and it was okay.

Do you know where i can find a simple procedure for to measure with an fluke 45.

I want to measure resistance off 100ohm. I search on the procedure site but i don't find procedure with an fluke 45.

Thanks

0
Avatar
Dale Chaudiere

Where did you search.  We have a number of Fluke Calibration warranted verification and adjustment procedures for the 45, RS-232 and IEEE-488 closed-loop.  If you are using the 45 as reference, use the 45 FSC.

0
Avatar
David Van der Meersch

Hi Michael,

 

Where can i find an acc file for fluke 45 for 12 months?

 

 

0
Avatar
David Van der Meersch

Hi Michael,

 

Where can i find an acc file for fluke 45 for 12 months?

 

 

0
Avatar
Dale Chaudiere

Copy 45.acc to alt\45_6mo_99pct.acc

Edit 45.acc:

Change "6 months" to "12 months" in header.

Added 0.005 to each "Tol." entry in

Mode: Volts

# DC

e.g.   0.02 => 0.025

That's it.  DC Voltage is the only 45 specification that has "6 Months" and "1 Year" entries.

Reference:

FLUKE 45, Dual Display Multimeter, Users Manual, PN 855981, January 1989, Rev. 4, 7.97

Please sign in to leave a comment.