Agilent 34450A Calibration Procedure

Hi anybody have calibration procedure for Agilent 34450A ? manual or by Visa connection

thank you 

4 comments

Date Votes
0
Avatar
Kenneth Grubaugh

I'm having issues writing a procedure in Met Cal Editor. The problem is with the Visa command CALC:STAT ON which throws an error on the Keysight 34450A. This is for the null operation required for the zero verification. I start with:

CALC:FUNC NULL

CALC:STAT ON

When I use National Instruments Measurement & Automation Explorer VISA Test panel these commands work perfectly, but stepping through debug operation in Editor no such luck.

If I can get past this issue, I will have the procedure complete. Does anyone have any suggestions?

0
Avatar
Jerral Plowman

That's a common command, the only thing I can think of is check the line number in Editor and make sure they are not numbered the same.

Example

1.001 VISA         CALC:STAT ON

1.001 VISA         CALC:FUNC NULL

 

0
Avatar
Kenneth Grubaugh

You are correct that was my problem. Thank you for your help Jerral.

 

If anyone is interested in getting this procedure, just respond to this thread with email and I will send it to you.