I'm using the NA2250 to obtain a VRMS measurement of 300V 400Hz. I'm able to tell the unit what settings I want but, I'm unable to obtain the measured value using [I] or [I$] (currently using the MEMI command to input the display data).

I’m using METCAL V7.2  and failed with this attempt (and a few different combinations).

  IEEE         [D2000]
 IEEE         "INX VRMS"[13][10]
  IEEE         [D2000]
  IEEE         "FTH VRMS"[13][10]
  IEEE        [I]
  MEMCX   300   V             2U          400H

If anyone could offer working data retrieval commands? I would be very grateful.

Mark

 

5 comments

Date Votes
0
Avatar
Mike McGrew

You need to use INX MAIN to read the main display or INX FREQ to read the frequency display.

Try this:

 

IEEE     FNC RMSV[X0D0A]                                ;Configure Total mode

IEEE     SRX VOLT 500[X0D0A]                          ;Configure for 500 volt range

IEEE     INX MAIN[X0D0A][D500][I]                     ;Initiate read of main display

 

This code works for me.  Hope it helps.

 

Mike

 
0
Avatar
Mark Evans

Hi Mike, thanks for your reply.
I’m unable to use the Hex code (illegal command).

The 2250 will only respond when I use " "

This attempt hangs at step 1.009 (or 1.008 if [I] is on same code line)
# set up 2250
  1.001  IEEE         "FNC RMSV"[13][10]
  1.002  IEEE         "SRX VOLT 500"[13][10]
  1.003  IEEE         "SET RREF OFF"[13][10]
  1.004  IEEE         "SET VARI OFF"[13][10]
  1.005  IEEE         "SET RTIO OFF"[13][10]
  1.006  IEEE         "SET ACAL ON"[13][10]

Apply 300V 400Hz, wait for lock then advance

# get display value
  1.007  IEEE         [D1000]
  1.008  IEEE         "INX MAIN"[13][10]
  1.009  IEEE         [D500][I]
  1.010  MEMCX  500   300.0V         2U          400H

Any suggestions welcome

Mark

0
Avatar
Mike McGrew

Im not using the quotation marks in my code. My MET/CAL version is 8.1.7.

Try this:

IEEE   [TERM CR]

to specify the terminating character from the device.

0
Avatar
Mark Evans

Success.

Thanks for your help Mike.

Mark.

0
Avatar
Mike McGrew

Glad to help.