Issues with Fluke 5500A/COIL

Hello

I have some problems with an auxiliary device the "Fluke 5500A/COIL" it needs to be configured but when I try to do tis in MET/CAL Editor it display the following message " E2494: The System Configuration is in use by aother application ("Editor"),

How can I fix this?

How can I configure the 5500A/COIL

Please and thank you

3 comments

Date Votes
0
Avatar
William Green

The 5500A/COIL is configured under the 5500, 5502, 5520 or 5522 under Tools - Configure Instruments in Editor.

0
Avatar
William Green

Forgot to add, I don't think you can have it configured on two instruments.

0
Avatar
Michael Johnston

His issue isn't with the Coil config specifically, it's with the config file being locked.

You need to open your metcal.ini file (the easiest way to get to it is to go to your Start Menu>Programs>Fluke Calibration>Fluke Program Data). Toward the bottom of the file will be a line that says "lock=Editor" which you need to change to "lock=None"