How can use acc file?


Hello,

I need to help with the acc file. I´m trying to use the datas of the acc file in a procedure but it gives me the following errors: E136, E1896.

 I'd like to know if there is something that I have to configure. thanks

8 comments

Date Votes
0
Avatar
Chad D.

Hello ALHELI, can you post your procedure code?

0
Avatar
ALHELI MORENO ROMAN

Hello Chad;

Thanks for your help.

I tried to compile this:

1.001 STD  5522A

1.002 M[5]=320

1.003 M[6]=ACCV("5522A","330V",M[5])

But I couldn't.

 

 

 

0
Avatar
Chad D.

I will assume that your compiled procedure includes the MATH FSC. Per the MATH FSC help topic, to access accuracy files using the MATH FSC's ACCV() function, you must provide the following information:

acc = ACCV(device, mode, val)

where the three arguments are: String, String, Numeric

The first argument <device> is the device name. The device must be a configured instrument (system instrument*, user-configured instrument, or manual standard), and must match the instrument listed in the accuracy file's header verbatim.

The second argument is an accuracy file mode string. This string selects the appropriate section of the accuracy file, and must match the mode string within the accuracy file verbatim (e.g. use "Volts" to match Mode: Volts).

The third argument <val> is the specified lookup value. MET/CAL will look within the requested accuracy file mode for this value, which must fall between the values in the first and second column of the accuracy file.

Applying these instructions to your procedure code results in the following changes:

 

 1.001 STD Fluke 5522A 1.002 MATH M[5] = 320 1.003 MATH M[6] = ACCV("Fluke 5522A", "Volts", M[5])

 

However, I suggest using named variables instead of the numbered Memory Registers to improve code readability. For example:

 

 1.004 STD Fluke 5522A 1.007 MATH LookupValue = 320 1.008 MATH Accuracy = ACCV("Fluke 5522A", "Volts", LookupValue)

 

0
Avatar
ALHELI MORENO ROMAN

Hello Chad, thanks for your help. It was exactly what I needed.

0
Avatar
ALHELI MORENO ROMAN

Please, could you tell me how can I do a file for the uncertainties of the certificate of calibration?. Has Metcal an application for this?. Thank you.

0
Avatar
metrologia

Can anyone tell me the proper commands to see Unit VAR in Test Report when I use the following procedure?


 10.001  HEAD         {Accuracy Power 500 VAR}
 10.002  DISP         Select on 5520A 20A.
 10.003  M5520        5.0A           90deg                        SI HC
 10.004  5520         100.00V                      60H            SI    S  2W
 10.005  MATH         L[1] = ACCV2("Fluke 5520A", "Amps SI HC", 5.0, 100)
# Convert Amps to percent
 10.006  MATH         L[2] = (L[1]/5.0) * 100
 10.007  MATH         L[1] = ACCV2("Fluke 5520A", "Volts SI", 100, 100)
#Convert Volts to Percent
 10.008  MATH         L[3] = (L[1]/100)* 100
# RSS the volts and current together per 5520A manual. This assumes a PF=0
 10.009  MATH         L[6] = RSS_L (2,3)
# Use the result to set the accuracy
 10.010  ACC          500VAR         L6%
 10.011  MEMI         Enter reading in VAR
# UUT Tolerance
 10.012  TOL          0.2%
 10.013  MEMCX  500   500.000'VAR'   TOL
 
I only can see the value but no unit ( W no problem)
 

Regards

0
Avatar
metrologia

How can I test i430flex test points are: 3000A- 2000A-1000A with 5520A and coil 150 turns.

 