I am new to writing procedures and was wondering if someone could explain why the following code fails with an error stating that the tolorance cannot be calculated.

1.023   TOL    .103%  .1/  .001U

1.024   MEMCX  15  7.00000V   TOL    60H

If I modify the TOL line to just one value   (1.023   TOL    .103%)  it will calculate the tolerance correctly. But as soon as I add another value it errors.

The TOL help information shows the TOL function with multiple adders. Why won't this work?

Thanks

 

6 comments

Date Votes
0
Avatar
Nathan Ryder

Can you give us the code leading up to this?

 

If I do this

5502E +Voltage = 7V

MEMI Enter UUT reading

TOL  .103%  .1/  .001U

MEMCX 15 7.00000V TOL 60H

It works fine

Also what version are you running?

0
Avatar
Bill Sierschula
Here is the code.  Also we have version 8.1.3
 
Thanks 
 
1.001  ASK-   R     N              P                            A        V
  1.002  ASK+           K
 
  1.003  MATH         @ReqMetCalVersion = "7.20x"
  1.004  CALL         Sub MET/CAL Version Check
 
  1.005  IF           NOT(@MetCalVersionOK)
  1.006  DISP         This procedure requires MET/CAL or 5500/CAL 7.20 SP1
  1.006  DISP         (7.20x) or later.
  1.007  END
  1.008  ENDIF
  1.009  CALL         Sub Display Copyright
  1.010  CALL         Sub High Voltage Safety Warning
 
  1.011  HEAD         TEST CONDITIONS
  1.012  DISP         - No specific warm-up time requirement.
  1.012  DISP         - Ambient temperature: 18°C - 28°C.
  1.012  DISP         - Relative humidity: less than 90%.
 
  1.013  HEAD         INITIAL INFORMATION
  1.014  CALL         Sub Select Failed Test Mode
  1.015  CALL         Sub Select Verification Method (No NTHROW)
 
  1.016  JMPL         END                          @Abort
 
  1.017  RSLT         =
  1.018  HEAD         {VOLTAGE TEST}
 
  1.019  TARGET       -p
#  1.043  HEAD         -2 [7][D250][7][D250][7][D250][7][D250][7][D250]
#  1.044  HEAD         -2 100 mV Range
#  1.045  SCPI         *RST
  1.020  DISP         Make the following connections:
  1.020  DISP         [32]          5520A             to     UUT FRONT
  1.020  DISP         [32] NORMAL V/Ohm/Diode/RTD HI ---> Input V HI
  1.020  DISP         [32] NORMAL V/Ohm/Diode/RTD LO ---> Input V LO
  1.020  DISP         [32]  AUX A/Ohm-SENSE/AUX V HI ---> Input A HI
  1.020  DISP         [32]  AUX A/Ohm-SENSE/AUX V LO ---> Input A LO
 
#  7.009  SCPI         :SEL:VLT
#  7.010  SCPI         VOLT:AC:RANG 100mV;RES 0.0001mV
  1.021  5520         7.00000V                     60H            SI    S  2W
  1.022  TARGET       -m
#  7.013  SCPI         INIT;*TRG;FETCH?[I]
  1.023  TOL          .103% .1/ .001U
  1.024  MEMCX  15    7.00000V       TOL           60H

 

0
Avatar
Nathan Ryder

Can you generate a T.U.R report for it?........In editor under "Tools" then "Generate T.U.R report.

0
Avatar
Bill Sierschula

It generates the following response...

 

1.023  TOL   .130%  .1/  .001U

1.024  MEMCX 15  7.0000V  TOL  60 H

#!  T.U.R. not calculated because tolerance depends on numeric register.

 

 

0
Avatar
Michael J.

Hi,

Try adding leading zeros to each component. It looks like it's interpreting at least one of them as a variable name.

0
Avatar
Bill Sierschula

We figured it out. It looks like there can only be one space between the variables. It worked with or without the leading 0's