8846 time out

Since updateing to metcal 8.3 all programs that have an 8846A as the standard now time out, program puts the meter in the right mode/range but does not read back the value and time out. Procedures worked before and tested same procedure susbstituting an 8508A for the 8846A and the procedure worked .  Is this a problem with the 8846A firmware or with MetCal 8.3? I have tried several 8846A meters on the bench.

 

Brian Knoblauch

NIAR Metrology

3 comments

Date Votes
0
Avatar
William (Bill) Spath
Bug in v8.3. Download and install Update 1 for v8.3
0
Avatar
Brian Knoblauch

Where is update 1.  I do not see it on the Gold member Met/Support website.

0
Avatar
Joakim Spångberg

Are you looking at the new support website? flukecal.com -> Service and Support -> My MET/SUPPORT. I am not sure if they update the old one since the new one came out of beta.