Hi

if i need withdarw data by ieee from 2 device , but at the same time, how can i do it by met/cal 8.2? 

for example i have two DMM (34401A) and i want to take measuring reading of both at the same time and this point very important. t

thank you friends

6 comments

Date Votes
0
Avatar
Dale Chaudiere

Instrument Names (System Instruments, User-configured Instruments, and Manual Standards) are unique in MET/CAL.  Only one unique Instrument Name (e.g. "HP 34401A") can be configured at any time.  The 34401 FSC System Instrument name is "HP 34401A".  The 34401 FSC can be used to control only one 34401A.

Here is what I would do.

1. 1st 34401A: Configure "HP 34401A" and use the 34401 FSC or DMM Flexible Driver with alias "DMM".

2. 2nd 34401A: Configure "Agilent 34401A" with alias "DMM2".  Clone the DMM Flexible Driver as follows:

2a. Copy sub_driver_dmm.txt to sub_driver_dmm2.mc

2b. Copy sub_send_cmd_dmm.txt to sub_send_cmd_dmm2.mc

2c. Change ALL instances of [@DMM] in the above cloned files to [@DMM2].

2d. Call the subprocedures in sub_driver_dmm2.mc and sub_send_cmd_dmm2.mc to control the second 34401A.

0
Avatar
Denis Malakhov

i can't understand nothing. what is ub_driver_dmm.txt . can you send one txt file for example or explain more deep? thank you very much

0
Avatar
Dale Chaudiere

If you are a MET/SUPPORT Gold subscriber, the file (sub_driver_dmm.txt) is included with any Fluke Calibration warranted procedure with dmm or DMM in the file name and /DMM in the INSTRUMENT lines.

0
Avatar
Denis Malakhov

ok. i will find. thank you

0
Avatar
Denis Malakhov

if i can't find sub_driver_dmm.txt and i am met/cal gold member, what can i do more? where i can find it?

thank you

0
Avatar
Dale Chaudiere

Did you search for and download a procedure with "DMM" in the project name?  Projects for main procedures with "DMM" in the name will include the subprocedures for the DMM Flexible Driver.