Hi Harri,
Are you using MET/CAL for these procedures? If so, leaving the barcode field blank in the Instrument Configuration screen will cause the barcode to be prompted for each of those instruments when you start the procedure.
Hello.
We calibrate Agilent E4980A using Agilent reference capacitors and resistors and HP 3458A.
We have six pieces of HP 3458A, two sets of reference capacitors and one set of resistors.
In calibration we use MET/TEAM procedure and only one piece of each standard.
Is it possible to make a calibration procedure that asks from the calibration technician what individual HP 3458A and capacitors and resistors will be used in this calibration?
Then STDs traceability would be OK.
Kind Regards,
Harri Aarrelampi
Finland
Picture where the STDs are located for every calibration.
Hi Harri,
Are you using MET/CAL for these procedures? If so, leaving the barcode field blank in the Instrument Configuration screen will cause the barcode to be prompted for each of those instruments when you start the procedure.
Interesting. There is no refer to that in the manual. We will test this and I try to remember to report the result here. Thank you very much for your answer. Merry Christmas and Happy New Year to Fluke from Finland.
It seems to be working. Simple. Now we create "universal assets" and refer to the actual assets when performing calibrations. We have to implement this slowly and test other procedures.